Quick start - F620
MR-28300A-U Quick start manual ATEQ F620 Page 2/24
PREAMBULE
1. DEFINITION OF THE ATEQ F620
The ATEQ F620 is a compact air/air
leak detector used to test the air-
tightness of parts. The method used
is based on the measurement of a
small variation or drop in differential
pressure between the test and
reference parts, when both are filled
to an identical pressure.
TEST
PART
REFERENCE
PART
DIFFERENTIAL
SENSOR
PIEZO
SENSOR
TEST
PRESSURE
2. MEASUREMENT CHARACTERISTICS
2.1. PRESSURE DROP MEASUREMENT
RANGE ACCURACY
RESOLUTION
Maximum
0 – 50 Pa +/- (2,5% of the pressure + 1 Pa) 0,01 Pa
0 – 500 Pa +/- (2,5% of the pressure + 1 Pa) 0,1 Pa
0 – 5000 Pa +/- (2,5% of the pressure + 10 Pa) 1 Pa
2.2. TEST PRESSURE MEASUREMENT
RANGE ACCURACY
RESOLUTION
Maximum
F.S. = 75 mbar* +/- (1,5% of the pressure+ 0,2 hPa) 0,1 % F.S.
F.S. < 0,3 bar +/- (1,5% of the pressure + 1 hPa) 0,1 % F.S.
0,3 F.S. 1 bar
+/- (1,5% of the pressure + 3 hPa) 0,1 % F.S.
1 < F.S. 5 bar
+/- (1,5% of the pressure + 7.5 hPa) 0,1 % F.S.
5 < F.S. 10 bar
+/- (1,5% of the pressure + 15 hPa) 0,1 % F.S.
10 < F.S. 20 bar
+/- (1,5% of the pressure + 30 hPa) 0,1 % F.S.
* Specific (relative)
F.S. = Full scale.