Chapter 4 – Functions of the instrument
UM-23100B-U User manual ATEQ F CLASS Page 68/90
Press the START key to start
the learning cycle. At the end
of the cycle, the readout
displays the test results in the
calibration unit.
CYCLE/Pr :001
PRESS = 0.505 bar
LEAK = 1.5 Cal-Pa
READY
To stop the cycle, press the
RESET key.
CYCLE/Pr :001
PRESS = 355.5 mbar
READY
Note: the CDF (leak flow calibrator) can be used to calibrate the instrument and to send
this value via the infra-red link.
3.7.2. CAL check
This special cycle is used to verify calibration in calibrated mode. See explanation in
previous paragraph. The CAL check cycle checks whether the calibration has drifted
beyond the limits set as a percentage. If these have been exceeded, an alarm will be
triggered and a calibration cycle or an instrument check will be required.
If the opposite is the case the "Pass" and "end of cycle" or "Fail" and "end of cycle"
outputs are activated depending on the measurement made in comparison with the
reject level.
The CAL check request may be made from input 7 on connector J3 programmed for this
function, or from a programmed function key on the optional RC5 keypad, if this is
installed.
In the main menu, position the
cursor beside SPE CYCLE
and confirm using ENTER.