Vertical Laminar Flow Cabinet
BBS-V1300/V1800
User Manual
BIOBASE GROUP
Version 2021.05
Question and Answer IconNeed help?

Do you have a question about the Biobase BBS-V1300 and is the answer not in the manual?

Summary

Unpacking, Installation and Debugging

Unpacking Procedures

Procedures and tools for safely removing the cabinet from its packaging.

Accessories Checking

Verifying all included components against the packing list for completeness.

Installation Conditions and Operating Environment

Specifies the required environmental parameters for cabinet installation.

Installation Steps

Step-by-step instructions for assembling and setting up the cabinet.

Post-Installation Inspection

Procedures to verify the cabinet is installed and functioning correctly.

User Instructions

Product Functions and Concepts

Overview of the cabinet's capabilities, purpose, and airflow mechanisms.

Product Structure Details

Details the physical components, layout, and individual parts of the cabinet.

Operation Instructions

Guidelines and step-by-step procedures for operating the cabinet safely and effectively.

Daily Maintenance Procedures

Routine cleaning and maintenance tasks for the cabinet's operation and longevity.

Replacement Parts List

Lists the part numbers and specifications for replaceable components for the models.

Wiring Diagram

Electrical schematic illustrating the internal wiring connections of the cabinet.

Trouble Shooting and Solution

Common Faults and Troubleshooting Steps

Identifies common operational faults and provides step-by-step solutions.

Fuse Replacement

Instructions on how to safely replace the fuse in the unit.

Fluorescent Lamp Replacement

Steps for replacing the fluorescent lighting component.

UV Lamp Replacement

Procedure for replacing the UV lamp tube and resetting its timer.

Label Descriptions

Explains the meaning of various labels and symbols found on the equipment.

Warranty Information

Warranty Terms and Conditions

Details the warranty period, terms, conditions, and product lifespan.

Overview

This document is the user manual for the BIOBASE Vertical Laminar Flow Cabinet models BBS-V1300 and BBS-V1800.

Function Description

The BIOBASE Vertical Laminar Flow Cabinet (LAF) is a workbench or similar enclosure designed to create a particle-free working environment. Its primary function is sample protection by taking air through a filtration system and exhausting it across a work surface in a laminar or unidirectional air stream. The cabinet is enclosed on the sides and maintained under constant positive pressure to prevent the infiltration of contaminated room air. It is used to protect experimental materials and create a local high-cleanliness air environment, ensuring the accuracy of experiments within the operating area. However, it does not provide protection for the environment or operators.

The cabinet features a driving system for the front window, consisting of a tubular motor, the front window itself, a hauling mechanism, and a limit switch. The air filtration system is a core component, comprising a blower and a HEPA filter system. This system transfers filtered air to the work area, ensuring airflow velocity and maintaining Class 100 cleanliness. For disinfection, a UV lamp is located at the top of the work zone, emitting a wavelength of 253.7nm to effectively disinfect the entire area. Illumination is provided by an LED light, ensuring the standard requirement for average illumination is met.

Important Technical Specifications

The manual provides technical parameters for two models: BBS-V1300 and BBS-V1800.

External Size (WDH):

  • BBS-V1300: 13107502040 mm
  • BBS-V1800: 18007502040 mm

Working Zone Size (WDH):

  • BBS-V1300: 1200645610 mm
  • BBS-V1800: 1700645610 mm

Power Supply AC:

  • BBS-V1300: 220V±10%
  • BBS-V1800: 110V±10%

Frequency:

  • BBS-V1300: 50 Hz
  • BBS-V1800: 60 Hz

Consumption:

  • BBS-V1300: 400W
  • BBS-V1800: 450W

Airflow Velocity: 0.3~0.5m/s for both models.

UV Lamp Consumption:

  • BBS-V1300: 30W*1
  • BBS-V1800: 40W*1

Fluorescent Lamp Consumption:

  • BBS-V1300: 12W(LED)*1
  • BBS-V1800: 16W(LED)*1

HEPA Filter: 99.999% efficiency at 0.3µm for both models.

Noise: <65dB for both models.

Additional Notes on Electrical Properties:

  • The cabinet should not break down in 5s when the voltage increases by 1390V (AC) within 5s.
  • Grounding resistance: ≤0.1Ω.

Working Environment:

  • Only applicable for indoor operation.
  • Ambient temperature: 15°C ~35°C.
  • Relative Humidity: ≤75%.
  • Atmospheric pressure range: 70KPa~106KPa.
  • Adequate power supply to the laminar flow cabinet.
  • Power supply needs to be grounded (voltage between neutral and ground should equal to 0).

Usage Features

Installation: The cabinet should be placed in a protected area, away from doors, windows, and air conditioning outlets to prevent airflow impact. A minimum 300mm gap must be maintained on the side and back for clean operation and inspection. The base stand is assembled with the main body using hexagon socket head cap screws. The foot height can be adjusted; clockwise rotation lowers the feet, allowing movement, while anticlockwise rotation raises them, fixing the cabinet in place. A gas (water) tap can be installed, with its mouth inside the operating zone. After installation, verify voltage and frequency match the nameplate, and check fan motor, fluorescent lamp, UV lamp, screen buttons, and socket functionality.

Control Panel: The control panel features soft-touch buttons and an LCD display.

  • Power Button: Switches the machine between suspension and working modes.
  • Blower Button: Controls blower working status. It has a memory function for the last used gear. (Does not work when the front window is fully closed).
  • Fluorescent Lamp Button: Controls the fluorescent lamp.
  • UV Lamp Button: Controls the UV lamp. This function is interlocked with the blower, fluorescent lamp, and front window; it will not work if the fluorescent lamp, blower, or front window are open.
  • Socket Button: Controls the power status of the waterproof socket.
  • Glass Window Up/Down Buttons: Control the motorized front window. Pressing causes a sound; holding moves the window continuously.
  • Air Volume Increase/Decrease Buttons: Adjust blower speed. Each press changes the gear and rings a buzzer. These buttons also adjust UV lamp timing (10-90 minutes).
  • LCD Display: Shows blower wind speed strength, current wind state, time, UV lamp working time, and HEPA filter working time.
  • Clock Adjustment: In standby mode, press the light button continuously to enter clock setting mode.
  • UV Lamp Timing: In standby mode, when the front window is lowered to a minimum, press the UV lamp control button to open the UV lamp. Timing can be adjusted using the air volume buttons.

Waterproof Socket: Located in the operating area, it provides power to equipment. The total load of sockets should not exceed 500W. The socket is waterproof only when its front cover is down.

Normal Operation Notice:

  • Ensure correct and stable input voltage, with the main power socket's rated load higher than cabinet consumption. The plug must be well grounded.
  • When moving multiple samples, move low-polluting samples to high-polluting samples slowly and stably.
  • Item weight in the cabinet should not exceed 23Kg/25×25cm².
  • Avoid vibration from equipment like centrifuges or vortex oscillators, as this can cause contamination from the HEPA filter.
  • Do not use open flames, as they disrupt airflow. Infrared sterilizers are recommended for sterilization.
  • HEPA filter life: Replace when resistance increases to the point where speed requirements cannot be met. Used filters should be processed as medical waste.
  • Do not remove or loosen screws of parts.
  • Maximum storage period is one year; performance tests are required if stored longer.
  • UV lamp function is only available when the front window is fully closed.
  • Sterilize the cabinet with the UV lamp for at least 30 minutes with the window fully closed before any experiment.
  • For eye and skin safety, people should leave the room during UV sterilization.
  • UV lamp should be checked regularly and replaced after 600 hours of total working time or if intensity is low.
  • Move the front window to a suitable height and turn on the fan. Start experiments after the fan has worked for at least 30 minutes.
  • For operating safety, place experiment materials inside the cabinet before starting.
  • After experiments, fully close the front window and sterilize with the UV lamp for 30 minutes before turning off the cabinet.

Maintenance Features

Daily Maintenance:

  • Disinfect and clean the operating area.
  • Clean the external surface and front window around the operating area using a soft cotton cloth with concentrated liquid soap, then clean water, and finally dry. For contaminated surfaces, use 70% rubbing alcohol or other disinfectants (that do not damage 304 stainless steel).
  • Check various equipment functions.
  • Record maintenance results.

Monthly Maintenance:

  • Clean external surface and front window.
  • Use a towel with 70% rubbing alcohol or 1:100 dilution of household bleach (0.05% sodium hypochlorite) to wipe the working table, inner face of the front window, and inner wall surface of the working area (excluding the top wind grid). Use another towel with sterile water to remove chlorine residue.
  • Check various functions of the cabinet.
  • Record maintenance results.

Annual Maintenance:

  • Check the two lifting belts of the front window tubular motor for proper connection and tightness.
  • Check the UV lamp and LED light.
  • Apply for an overall performance test annually to ensure safety requirements are met. (User is responsible for testing costs).
  • Record maintenance results.

Replacement Parts: The manual includes a list of replacement parts for both BBS-V1300 and BBS-V1800 models, including fuses, lamp holders, UV lamps, LED holders, UV lamp ballasts, HEPA filters, fans, control panels, front window glass, side window glass, and single wall gas taps.

Troubleshooting: A troubleshooting guide is provided for common issues like fluorescent lamp not working, UV lamp not working, buttons not working, and no electricity in the equipment. It lists check parts and corresponding measures.

Fuse Replacement: The fuse of the waterproof socket is on the right side of the cabinet. To replace, turn off power, disconnect the plug, use a flathead screwdriver to unscrew the fuse holder, replace the fuse, and reassemble.

Fluorescent Light Replacement: Turn off power, remove the LED stand, unplug the right side, replace with a new LED stand, and insert into the inclined slot.

UV Lamp Replacement: UV lamp tubes should be changed every 600 hours. Turn off power, disconnect the plug. Rotate the tube 90° to remove it, then install a new one by rotating in reverse. After replacement, reset the accumulated using time to zero by pressing and holding the UV button for 5 seconds until a buzz alarm sounds.

Storage Conditions: Laminar flow cabinets should be stored in a warehouse with relative humidity no more than 75% and temperature below 40°C. The warehouse must have good ventilation and be free of acid, alkali, or other corrosive gases. Storage period should not exceed one year. If stored for more than one year, the cabinet needs to be unpacked and checked before selling and using. Only tested and qualified safety cabinets should be sold.

Warranty: The warranty is 12 months from the EX-factory date, excluding consumable accessories (UV lamp, fluorescent lamp, and fuse). Biobase is not liable for damage caused by improper operation. Expired warranty repairs will incur charges. The lifetime of the laminar flow cabinet is 8 years from the production date. Biobase will provide equipment drawings and technical data for maintenance companies or personnel trained by Biobase engineers. The warranty declaration is: One-year Warranty, Life-long Maintenance.