VI. SampleLoading(cryo)
1. Makesuretheanticontaminatordewarisfilledwithliquidnitrogen(LN2).
2. The cryo holder should be pumped at least 2hrs in the Dry Pumping Station (see separate
protocol)beforeuse.Preferably,you'llhavepumpedtheholderthedaybeforeforatleast6hrs.
3. Place the cryo transfer station on the table top. Insert the cryo holder into the station and
removetheclipringandpreviousgridifpresent.
4. Make sure you have gathered your cryo sample in transfer dewar, forceps, large forceps for
gridboxtransfer,andanythingelseneededandplaceitnearthetransferstation.
5. Add LN2 to both the cryo transfer station and the cryoholder. These will boil at different times.
Do not overfill the holder nor the station, this will only cause the holder to become icy and the
transferstationtofrost.
6. Pluginthecontrolunitcordandturnonthetemperaturecontrolunit.
7. Tilt the compustage to 60° by going to the Search tab, flapout window, Control tab, Alpha
toggle,typein60°andclickSetAlpha.TurntheTurboon.
8. Check that the airlock pump time is set to 6090s (Setup tab, flapout window, Settings tab,
Defaultairlocktime:60s).