EasyManua.ls Logo

EDAN F6 - Page 152

EDAN F6
194 pages
To Next Page IconTo Next Page
To Next Page IconTo Next Page
To Previous Page IconTo Previous Page
To Previous Page IconTo Previous Page
Loading...
F Series Fetal & Maternal Monitor User Manual Product Specifications
- 141 -
I
sppa.3
<190W/cm
2
I
spta.3
<94mW/cm
2
Max Output Power <15mW
Effective Radiating Area: 942 mm
2
± 15%
FHR Measurement Range: 50 bpm ~ 240 bpm
Resolution: 1 bpm
Accuracy:
±2 bpm (F6/F6 Express)
±1 bpm (F9/F9 Express)
Dielectric Strength: 4000Vrms
Maximum transducer
temperature rise during
use:
Less than 5 qC (9ºF)
Expanded uncertainty of
temperature test:
U=0.4qC (0.72ºF), k=2
DECG
Technique: Peak-peak detection technique
DFHR Measurement
Range:
30bpm ~ 240bpm
Resolution: 1bpm
Accuracy: ±1bpm
Input Impedance: > 10M (Differential, DC50/60Hz)
Input Impedance: > 20M (Common Mode)
CMRR: > 110dB
Noise: < 4μVp
Skin Voltage Tolerance: ±500mV
Fetal Input Voltage
Current:
20μVp-3mVp
TOCO
TOCO Range: 0% ~ 100%
Non-linear Error: ±10%
Baseline Drift due to
Temperature Changes
1 unit/min/°C (free air)
5 units/min/°C (underwater)
Resolution: 1%
Zero Mode:
Automatic(TOCO value becomes zero or below
lasting for 30 seconds)/ Manual
Dielectric Strength: 4000Vrms
IUP
Pressure Range: 0 ~ 100mmHg
Sensitivity: 5μV/V/mmHg

Table of Contents

Related product manuals