F Series Fetal & Maternal Monitor User Manual Product Specifications
- 142 -
Non-linear Error: ± 3mmHg
Resolution: 1%
Zero Mode: Manual
Fetal Movement
Technique: Pulsed Doppler ultrasound
FM Mode Automatic/Manual
AFM Mode Trace or blackmark
Display
Range
0-999
ECG
HR Measurement Range: 30 bpm ~ 240 bpm
Input Signal Range: ±8 mV PP
HR Measuring Accuracy: ±2 bpm
ECG Waveform: Manual control ECG waveform display
ECG falls off: Detect automatically
Patient Leakage Current
(Limit)
N.C. S.F.C.
d.c. 10μA 50μA
a.c. 10μA 50μA
Patient Auxiliary Current
(Limit)
N.C. S.F.C.
d.c. 10μA 50μA
a.c. 10μA 50μA
Differential Input
Impedance
>5M
Display Sensitivity
2.5mm/mV (×0.25), 5mm/mV (×0.5), 10mm/mV
(×1), 20mm/mV (×2), AUTO gain
Sweep speed 25mm/s
d.c. Offset Voltage
Tolerance
±500mV
Auxiliary Current (Leads
off detection)
Active electrode: < 100 nA
Reference electrode: < 900 nA