  • M
    michaelcontrerasAug 16, 2025
    What to do if my EDAN F6 Medical Equipment has no display?
    • J
      jorozcoAug 16, 2025
      If your EDAN Medical Equipment has no display, the issue might be due to a loose power cable. Ensure the power cable is securely tightened. Another possibility is a blown fuse, which would require replacing the fuse. Also, check if the battery has run out of power; if so, connect the equipment to an AC power supply.
  • W
    wwashingtonAug 22, 2025
    How to fix a recorder error on EDAN F6 Medical Equipment?
    • N
      Nicole DavisAug 22, 2025
      To fix a recorder error on EDAN Medical Equipment, first ensure that the paper is loaded correctly and kept away from moisture. If the recorder hasn't been started, press the PRINT key. Check if the device has run out of paper and load more if needed. Also, verify that the paper drawer is completely closed and both latches are locked.
  • N
    Nicholas DavisAug 26, 2025
    What to do if my EDAN F6 shows no NIBP and SpO2 results?
    • J
      jmitchellAug 26, 2025
      If your EDAN Medical Equipment isn't showing NIBP and SpO2 results, check that the NIBP cuff is correctly wrapped around the patient's arm. If the NIBP cuff isn't inflating, extend the catheter and check its connection. Ensure the hose connector plug is securely connected to the NIBP socket and that the SpO2 sensor is properly connected to its socket. If the issue persists, try shutting off the power and then switching it back on.
  • Z
    zandrewsAug 28, 2025
    How to reduce noise from EDAN F6?
    • L
      Lee CarrilloAug 28, 2025
      To reduce noise from EDAN Medical Equipment, first, turn down the volume if it is set too high. If the noise is due to interference from a mobile phone or other electromagnetic source, turn off or move the source. Additionally, try moving the monitor to a location with less interference.
  • S
    Samantha WilkersonAug 31, 2025
    Why is there big ECG signal interference or a thick baseline on my EDAN F6 Medical Equipment?
    • J
      John MitchellAug 31, 2025
      Big ECG signal interference or a thick baseline on EDAN Medical Equipment might be due to abnormal electrode placement or invalid electrodes; check the placement and validity. Also, ensure the cable connector is well connected. Verify that the power socket has a standard ground wire and that the special ground wire connecting with the monitor is properly earthed.
  • B
    Brian PetersenSep 6, 2025
    What to do if my EDAN F6 has troubles with external contractions monitoring?
    • J
      josephhunterSep 6, 2025
      If your EDAN Medical Equipment experiences troubles with external contractions monitoring, first adjust the belt if it's too tight or too loose. If the belt has no elasticity, renew it. Maternal or frequent fetal movements can cause issues, so ask the patient to stay still or delay monitoring. If the body pressure from the uterus to the TOCO transducer is far higher than average, ensure favorable contact between the patient's skin and the TOCO transducer, and change its position if necessary.
  • H
    Harry ElliottSep 9, 2025
    What causes troubles with DECG FHR monitoring on EDAN F6 Medical Equipment?
    • D
      Diane ChristensenSep 9, 2025
      Troubles with DECG FHR monitoring on EDAN Medical Equipment can be caused by no ECG signal or bad contact between the reference electrode and the patient. In both cases, use a new spiral electrode.
  • Q
    qmeyerSep 12, 2025
    How to resolve issues with internal contractions monitoring on EDAN F6?
    • J
      Jason SteeleSep 12, 2025
      To resolve issues with internal contractions monitoring on EDAN Medical Equipment, consider these steps: * If the intrauterine catheter is jammed, wash it with a disinfectant. * In a “dry” environment or if the tip of the intrauterine catheter is placed extraovularly, wash with disinfector or change the position of transducer. * If the zero adjustment is wrong, zero the system. * If there is a connector failure, move or contact the catheter. If there is no trend fluctuation, change the intrauterine cable.
  • V
    Victor GrahamNov 18, 2025
    Why is my EDAN F6 Medical Equipment having trouble with ultrasound FHR monitoring?
    • F
      Frank MorrisNov 18, 2025
      Troubles with ultrasound FHR monitoring on your EDAN Medical Equipment can stem from several factors. If the patient is overweighted, consider monitoring FHR with DECG. Ensure proper ultrasound transducer position by adjusting it for a better signal. Also, check for a loose belt and tighten it if necessary. Wipe off any superfluous aquasonic coupling gel, or if there is inadequate gel, use the recommended quantity. Frequent fetal movements or maternal movement can interfere, so delay monitoring or ask the patient to stay still. Ensure you are not recording the maternal heart rate wrongly by changing the position of the ultrasound transducer. If the transducer is not well placed and mixed noise is recorded, adjust its position. Also, ensure you are using paper recommended by the manufacture...

Summary

Chapter 1 Safety Guide

Indications for Use

Specifies the intended applications and usage conditions for the fetal & maternal monitor.

Safe Operation Instructions

Provides essential guidelines for the safe operation of the fetal & maternal monitor.

Ultrasound Safety Guide

Details safety precautions specific to the use of ultrasound in fetal monitoring.

Safety Precautions

Outlines critical warnings and precautions to prevent injury or equipment damage.

Chapter 2 Installation Guide

Battery Installation

Step-by-step instructions for installing the rechargeable battery into the monitor.

Chapter 4 Alarms

Alarm Classification

Categorizes alarms into patient and technical types based on severity and source.

Alarm Treatment Measures

Provides guidance on responding to and managing alarms during monitoring.

Chapter 6 Pre-Monitoring Preparation

Chapter 7 Fetal Monitoring

Confirming Fetal Life

Explains methods to confirm fetal viability before starting monitoring.

Monitoring FHR with Ultrasound

Details the procedure for monitoring Fetal Heart Rate using an ultrasound transducer.

Monitoring FHR with DECG

Describes the process and precautions for monitoring Fetal Heart Rate using Direct ECG.

Monitoring Uterine Activity Externally

Details the procedure for monitoring uterine activity using an external TOCO transducer.

Monitoring Uterine Activity Internally

Explains the procedure for monitoring intrauterine pressure using an IUPC.

Start Monitoring

Outlines the steps to initiate the monitoring process on the device.

Chapter 8 Fetal Monitoring Display (F6;F9)

Chapter 9 Maternal Monitoring (F6 Express;F9 Express)

Maternal ECG Monitoring

Details the procedure for monitoring maternal ECG, including setup and operation.

Maternal SpO2 Monitoring

Explains how to perform maternal SpO2 monitoring, including sensor application and readings.

Maternal HR Monitoring

Covers maternal heart rate monitoring, including source selection and beep volume.

Maternal NIBP Monitoring

Describes the process for measuring maternal blood pressure using NIBP.

Maternal TEMP Monitoring

Details how to monitor maternal temperature using the TEMP sensor.

Chapter 10 Maternal Monitoring Display (F6 Express;F9 Express)

Chapter 11 FTS-3 Fetal Telemetry System

Installation Guide

Provides instructions for installing the FTS-3 fetal telemetry system.

Patient Application

Guides on applying the FTS-3 system for patient monitoring in various scenarios.

Chapter 12 After Monitoring

Chapter 13 Maintenance and Cleaning

Maintenance Procedures

Covers visual, routine, and mechanical inspections for equipment upkeep.

Chapter 14 Warranty and Service

Warranty Information

Details the terms and conditions of the product warranty provided by EDAN.

Contact Information

Provides contact details for technical support, service, and inquiries.

Appendix 3 Troubleshooting

No Display Issues

Lists common causes and solutions for the monitor not displaying any output.

Noise Issues

Addresses potential causes of noise in signals and their solutions.

Recorder Errors

Provides troubleshooting steps for common recorder malfunctions like paper jams.

Ultrasound FHR Monitoring Troubles

Offers solutions for issues related to inconstant traces or doubtful FHR readings.

DECG FHR Monitoring Troubles

Addresses problems with inconstant trends or displays during DECG monitoring.

Contractions Monitoring (External) Troubles

Solves issues related to bad trace quality or fluctuating TOCO baselines.

Contractions Monitoring (Internal) Troubles

Provides solutions for problems encountered during internal contraction monitoring.

ECG Signal Interference

Addresses causes and solutions for big ECG signal interference or thick baselines.

NIBP and SpO2 No Results Troubleshooting

Offers solutions for cases where NIBP and SpO2 measurements yield no results.

Blown Fuses

Guides on how to identify and replace blown fuses in the monitor.

Appendix 4 Ultrasound Intensity and Safety

Ultrasound Safety and ALARA Principle

Explains the principles of ultrasound safety, focusing on minimizing exposure.

EDAN F6 Specifications

General IconGeneral
SpO2 Measurement Range0 - 100%
Waterproof TransducerYes
PrintingBuilt-in thermal printer
Paper Speed1, 2, 3 cm/min
TypeFetal/Maternal Monitor
ParametersFHR, TOCO, ECG, SpO2, NIBP, TEMP
ECG Leads3-lead
Battery TypeRechargeable lithium-ion
Battery Life> 4 hours
TOCO Range0-100
AlarmAudible and visual alarms