General safety recommendations and repair instructions
7
6.3. Cleanliness during assembly
The greatest dangers to hermetically sealed compressors are dirt and moisture.
Particular care is therefore required when assembling such units:
Use only cooling coils (evaporators), condensers and tubing that are clean and dry on the
inside and have been delivered airtight.
Use only suitable refrigerant or nitrogen for pressure tests of those components.
Open the connections immediately before assembly, and never leave the tubing open longer
than necessary!
After cutting and deburring the tubing, carefully remove all metal shaving. Never use
compressed air.
Dry all components that have been exposed to air by heating and simultaneously evacuating
them, fill with inert gas and close.
Use a filter dryer not too small that has a close-meshed filter.
6.4. Leak detection and elimination
Wear a protecting mask to prevent poisoning by rests of gas!
Leak detection:
Measure the stand-still pressure using a manometer that is connected to the filling connector
(7/16").
Traces of the refrigerant in the system may be detected using an industry-standard electronic
leak detector.
In persistent cases, the refrigerant must be pumped out to a residual pressure of approx.
2 bar. The circuit is then to be subjected with dry nitrogen (humidity <2 v.p.m.) at a test
pressure of about 20 bar.
After leak detection has been completed, this mixture must again be pumped out from the
system.
Elimination:
Remove first the Schrader valve insert using a special tool.
To prevent moisture entering the system and scale formation during soldering, circulate a
forming gas (95 % nitrogen + 5 % hydrogen) through the system via the open filling connector
under moderate pressure.
Remove the filter dryer.
Solder the leak (using a soldering torch and silver solder S15). Wait until the soldering joint
has cooled down with the help of the circulating forming gas.
Insert a new filter dryer in the end of the capillary tubing and solder under the same conditions.
Perform a seal test of the refrigerant circuit. This means, the circuit is to be subjected with
nitrogen or refrigerant at pressure to detect leaks.
Screw in Schrader valve insert.
Evacuate the refrigerant circuit using a vacuum pump and a filling device for at least one hour.
Refill refrigerant.