Sheet #631u – Service special cycles
Version 1.04a  User guide ATEQ 6th series Page 5/5 
1.2.2.  Valve Auto-test 
The AUTO-TEST cycle must only be carried out with caps on the test and reference port. 
The test parameters are automatically configured according to the active program (test 
pressure) and the characteristics of the instrument.
The device returns to the "Cycle" menu and 
shows the special cycle selection. 
Plug the caps on the test and 
reference outputs. 
Press the Start to run the special cycle. 
52'%;%.'
AUTO TST 
Pr 001 
LEAK TEST 
Three complete cycles, Fill; Stabilization, Test 
and Dump are carried out. 
During the Auto-test, the device displays the 
measured pressure. 
FILL 
AUTO-TEST
Pr1 7.1s
The cycle stops automatically, if the valve is 
functioning properly, the instrument displays 
"TEST = PASS".
If the valve is leaking or with a fault, the 
following message appears: 
LEAK FAULT 
TEST = FAIL
600.0 
mbar 
OK
AUTO-TEST
Pr1 READY