Operating
T-SELECTIRTA-INTERNAL
It is entered into the circulator as the correction factor RTA
RTA-INT:
+01.56"C
and remains stored there.
RTA-NEW:
+02.78"C
The correction factor only refers to this one application. A
26.80 "C
i
new altered correction test setup. value is required for any new temperature or
a
RTA-EXT:
t02.89"C
RTA-NEW:
+01.87"C
26.80
"C
!
T-SELECTMIGH-LIMIT
1
H-LIMIT
+154"C
L-LIMIT:
-050°C
1
I
LL-NEW:
+3WC
/
26.90 "C
or:
H-LIMIT:
t154"C
L-LIMIT:
-027°C
I
LL-NEU:
-050°C
I
FUZZY
IDENT
43.20"C
Select T-SELECT
and
then RTA. Alter the display in
same way as described in chapter
13.4.
Only
if
a value input has been made it can be
switched over to negative
(-)
values.
13.6 Setting temperature limit values
The setting range of the operating temperature of the circula-
tor can be limited if the application or the flash point of the
selected heat transfer liquid requires this.
!
This is not a safety element but merely an aid to
help avoid
user
faults
when
operating the unit.
The
excess temperature protection must be set
sepa-
rately.
When in the external control mode (for EXTERNAL setting
see chapter
13.2),
the limits set restrict the temperature in
the circulator's own bath in order to guarantee
a
higher
degree of safety against unintentional heating up or cooling
down.
a
Select T-SELECT and then
LIMIT.
Alter the display in
same way as described in chapter
13.4.
First HIGH and then LOW-LIMIT.
If the HIGH-LIMIT value is lower than the set tempe-
rature value, the set value is decreased according
to the HIGH-LIMIT value.
13.7 Controlling heating and cooling
Heating and cooling are cycled.
!!!
is illuminated when
heating is activated.
@
or is illuminated when cooling
is activated. Flashing of the star means a cooling control bet-
ween
30
and
100%.
lights up if cooling is active with full cooling capacity,
X
lights up if cooling is active with partial cooling capacity,
8