Haake F6 - Sensor Breakage or Short Circuit; Undefined Fault; External Fault RS232 C

Haake F6
85 pages
Print Icon
To Next Page IconTo Next Page
To Next Page IconTo Next Page
To Previous Page IconTo Previous Page
To Previous Page IconTo Previous Page
Loading...
Fault
Displays
I
ALARM
/
15.4
Sensor breakage
or
short circuit
This fault can correspond to the internal sensor
(RF
internal) or to an external sensor
(RF
external) con-
nected to the socket on the rear.
I
ALARM
/
-
lmprove shielding of the sensor connector cable.
(
see chapter
17.5
)
The sensor must be tested and possibly
exchanged
by
qualified service personnel.
4320°C
The message
"REF
means that the reference resistor
for the measuring value evaluation is out of order.
-
lmprove shielding of the sensor connector cable.
(
see chapter
17.5
)
return unit for servicing.
15.5
Undefined fault
I
ALARM
I
This can be caused by fault which only occurs for
a
short period of time, i.e. with a fluctuating bath level
when the filling level is very close to minimum.
Before returning the unit, top up with heat transfer liq-
uid. This fault can often be remedied in this
way!
In all other cases this unit must be checked by qualified
service personnel.
15.6
External
fault
RS232C
L-----A
Check the external system.
ALARM
-.
RS
2326
The circulator has been switched to fault status via the
interface.