FIGURE 23 Error code
5. It indicates that the main unit operates normally when no error message
is displayed on the main interface. At this moment the respirator will not
execute any ventilation mode or parameter, but pop up a Select
Emergency Ventilation Mode dialog for your selection.
You can directly confirm the dialog and respirator will execute the default
ventilation mode and parameters.
Alternatively you can click on Ventilation Mode and select the desired
mode from the drop down list, and then set the related parameters in the
pop-up parameter setting window and save it, and respirator will execute
your selection.
A detailed description of the aforesaid operations is given in the
following section.
6. After setting respiration parameters and ventilation mode, make
connection to the patient.
7. During ventilation, adjust respiration parameters as necessary for the
patient.
Caution
Self-checking should not be a substitute for functional inspection. Before
using the equipment, functional inspection shall be performed in accordance
with Section 7 - Functional Inspection.
5.3
Select Ventilation Mode
To select ventilation mode, click on Ventilation Mode and select the desired
mode from the drop down list(As shown below).