Example
Oxygen Storage Capacity = 1000 L; MV = 11 L / min; oxygen concentration =
100%Calculated as follows
1000
L 100%
Ventilation Durationmin
=
×
= 91 min =1 h 31 min
11 L/min 100%
5.11
Alternative Breathing Apparatus
In case of any malfunction in T5 during respiration, use the following
alternative breathing apparatus:
Respiratory Leather Bag
1. Remove the patient breathing valve off the cannula or mask.
2. Connect respiratory bag and then perform manual respiration.
Failure of Oxygen
In case of failure of oxygen source for T5, the patient can inhale air through
patient breathing valve.
5.12
Battery Management
T5 is equipped with a built-in chargeable battery whose working time must
not be shorter than 5 hours in standard state.
The built-in battery should be recharged for not less than 8 hours, either
independently or by the main unit of T5 ventilator.
Battery Status Description
User can view in the main interface whether the battery is connected, being
recharged and the battery level, as detailed below
PARTS DESCRIPTION
1 indicates that the battery is not connected
2 indicates a battery level of 20 %
3 indicates a battery level of 40 %
4 indicates a battery level of 60 %
5 indicates a battery level of 80 %
6 indicates a battery level of 1000 %