H-1.601. 00137-B3.0
Page 50 of 81
• Example:
Oxygen Storage Capacity = 1000 L; MV = 11 L / min; oxygen concentration =
100%Calculated as follows:
Ventilation Duration(min)
=
×
= 91 min =1 h 31 min
11 L/min 100%
5.11
Alternative Breathing Apparatus
In case of any malfunction in T5 during respiration, use the following
alternative breathing apparatus:
• Respiratory Leather Bag
1. Remove the patient breathing valve off the cannula or mask.
2. Connect respiratory bag and then perform manual respiration.
• Failure of Oxygen
In case of failure of oxygen source for T5, the patient can inhale air through
patient breathing valve.
5.12
Battery Management
T5 is equipped with a built-in chargeable battery whose working time must
not be shorter than 5 hours in standard state.
The built-in battery should be recharged for not less than 8 hours, either
independently or by the main unit of T5 ventilator.
Battery Status Description:
User can view in the main interface whether the battery is connected, being
recharged and the battery level, as detailed below:
1 indicates that the battery is not connected
2 indicates a battery level of 20 %
3 indicates a battery level of 40 %
4 indicates a battery level of 60 %
5 indicates a battery level of 80 %
6 indicates a battery level of 1000 %