7 Functional Inspection
User must perform a functional inspection on the equipment prior to use
and after disassembly, or once every six months.
Tips:
When performing functional inspection on T5, the respiratory hose and
patient breathing valve must be connected.
In case that any malfunction or deviation to any set value is found in
functional inspection, the ventilator shall not be used until such malfunction
is remove in accordance with Section 8 - Trouble Shooting. If such
malfunction cannot be removed, contact Amoul or any professional
authorized by it for repair.
The full functional inspection shall include:
System Air-Tightness Inspection under 7.2
Examination of Patient Breathing Valve under 7.3
Machine Function Inspection under 7.4.
It is recommended to keep proper inventory of the following spare parts:
One-way diaphragm of patient breathing valve
7.1
Inspection Cycle
Prior to each use:
Perform a full functional inspection.
After use or disassembly:
Clean, disinfect or sterilize the equipment and its components (refer to
Section 6 - Sanitization);
Examine one-way diaphragm in patient breathing valve (refer to 7.3 -
Examination of Patient Breathing Valve). The diaphragm shall not be
wrinkled, become sticky or twisted.
Perform a full functional inspection.
As long as being idle
at least once every six months,Perform a full functional inspection.