11.5
Ventilation mode
Ventilation mode
Volume-control
mode
V-A/CV-SIMVIPPV
Pressure-control
mode
P-A/CP-SIMVCPAPPCV
Other mode
CPRManualHFNC
11.6
Supply specifications
Supply specifications
Air supply
Medical oxygen
Air supply pressure
2.76.0 bar
11.7
Ventilator specifications
Patient breathing valve resistance
Inspiration
6cmH2Oat flow rate of 3060 L/min
Expiration
6cmH2Oat flow rate of 3060 L/min
Emergency inhaled
air inlet
6cmH2Oat flow rate of 1530 L/min
Max. ventilation flow
Max. ventilation
flow
80L/min(@450kPa)
Activation Mode
Activation Mode
Pressure activation
Control Parameters
I/E
Adjustable between 9:1 and 1:9 with allowance of
±10%
Respiration
frequency
01~120bpm with allowance of ±1bpm or ±5%,
whichever is the larger
Tidal volume
50~2500mL (ATPD)
with allowance of ±30ml or ±15%, whichever is the
larger
Oxygen flow
velocity
180L/min with allowance of ±1L/min or ±15%,
whichever is the larger.
PEEP/CPAP
0,3~30cmH20 with allowance of ±2cmH20 or
±10%, whichever is the larger
Inspiratory pressure
5~60cmH20 with allowance of ±2cmH20 or ±10%,
whichever is the larger