H-1.601. 00137-B3.0
Page 67 of 81
11.5
Ventilation mode
11.6
Supply specifications
11.7
Ventilator specifications
Patient breathing valve resistance
<6cmH2Oat flow rate of 30、60 L/min时
<6cmH2Oat flow rate of 30、60 L/min时
Emergency inhaled
air inlet
<6cmH2Oat flow rate of 15、30 L/min时
Adjustable between 9:1 and 1:9 with allowance of
±10%
0,1~120bpm with allowance of ±1bpm or ±5%,
whichever is the larger
with allowance of ±30ml or ±15%, whichever is the
larger
1~80L/min with allowance of ±1L/min or ±15%,
whichever is the larger.
0,3~30cmH20 with allowance of ±2cmH20 or
±10%, whichever is the larger
5~60cmH20 with allowance of ±2cmH20 or ±10%,
whichever is the larger