EasyManua.ls Logo

Ambul T5 - Ventilation Mode; Supply Specifications; Ventilator Specifications

Ambul T5
95 pages
Print Icon
To Next Page IconTo Next Page
To Next Page IconTo Next Page
To Previous Page IconTo Previous Page
To Previous Page IconTo Previous Page
Loading...
H-1.601. 00137-B3.0
Page 67 of 81
11.5
Ventilation mode
Ventilation mode
Volume-control
mode
V-A/C、V-SIMV、IPPV
Pressure-control
mode
P-A/C、P-SIMV、CPAP、PCV
Other mode
CPR、Manual、HFNC
11.6
Supply specifications
Supply specifications
Air supply
Medical oxygen
Air supply pressure
2.7~6.0 bar
11.7
Ventilator specifications
Patient breathing valve resistance
Inspiration
<6cmH2Oat flow rate of 30、60 L/min时
Expiration
<6cmH2Oat flow rate of 30、60 L/min时
Emergency inhaled
air inlet
<6cmH2Oat flow rate of 15、30 L/min时
Max. ventilation flow
Max. ventilation
flow
≥80L/min(@450kPa)
Activation Mode
Activation Mode
Pressure activation
Control Parameters
I/E
Adjustable between 9:1 and 1:9 with allowance of
±10%
Respiration
frequency
0,1~120bpm with allowance of ±1bpm or ±5%,
whichever is the larger
Tidal volume
50~2500mL (ATPD)
with allowance of ±30ml or ±15%, whichever is the
larger
Oxygen flow
velocity
1~80L/min with allowance of ±1L/min or ±15%,
whichever is the larger.
PEEP/CPAP
0,3~30cmH20 with allowance of ±2cmH20 or
±10%, whichever is the larger
Inspiratory pressure
5~60cmH20 with allowance of ±2cmH20 or ±10%,
whichever is the larger