8 Trouble Shooting
If any failure cannot be removed, contact Amoul or any professional
authorized by it for repair. Stop using the machine in problem to avoid any
potential injury.
8.1
Technical Failure
Failure
Cause
Solution
T5 fails to start
T5 malfunctions
Contact the manufacturer or
your dealer for repair
Battery is exhausted
Recharge battery
Significant oxygen
loss
Air supply hose
leaks
Locate and eliminate
leakage
T5 cannot be shut
down
Improper operation
Long press Start/Shut Down
button for 3 seconds or
more
Power supply LED
operates unstably
Power supply plug
gets loose
Re-connect power supply
Short battery service
duration
Battery service life
expires
Replace battery with a new
one
8.2
Physiological Alarm
Message
Alarm
Cause
Solution
High
ventilation
rate per
min.
High
ventilation
rate per
min.
Exceed the set
upper limit of
ventilation rate per
min.
Examine the patient's
condition
Check whether the
upper limit is set
properly
Low
ventilation
rate per
min.
Low
ventilation
rate per
min.
Lower than the set
lower limit of
ventilation rate per
min.
Examine the patient's
condition
Check whether the
lower limit is set
properly
Asphyxia
Asphyxia
Asphyxia duration
exceeds the set
value
Examine the patient's
condition
Check whether the
duration value is set
properly
High airway
pressure
High airway
pressure
Exceeds the set
upper limit
Examine the patient's
condition