Operation of the equipment or system at a frequency lower than the
aforesaid range may cause incorrect results.
12.6
Basic EMC Properties of T5 Emergency
Ventilator
Basic EMC Properties of T5 Emergency Ventilator
T5 Emergency Ventilator will operate based on the parameters as set. For
more details see Section 11 in the manual. User can give alarm based on
real-time monitoring of operation state of T5 Emergency Ventilator, and
ensure accuracy of the following parameters in EMC environment
instructed for T5 Emergency Ventilator.
I/E
Adjustable between 9:1 and 1:9 with allowance of
±15%
Ventilation rate per
min. (MV)
1~45L/min. with allowance of ±0.5L/min or ±20%,
whichever is the larger
Airway pressure
monitoring
-20~100cmH20 with allowance of ±2cmH20 or
±15%, whichever is the larger
Activation pressure
-20~20cmH20 with allowance of ±1cmH20 or
±15%, whichever is the larger
Parameters of EMC Cable Material
Adapter input power
cable
0.8±0.01m
Adapter output
power cable
1.45+0.05m
Caution
EMC of the location where this environment is mounted and used shall be
adequately take into consideration in accordance with the said guidelines.
Any equipment on or near T5 ventilator may still cause interference with
T5 even though it is CISPR compliant, so user shall verify whether T5
operates normally before using it for patent.
Application of any unapproved part or component to this equipment may
impair its electromagnetic immunity and increase its EME.