H-1.601. 00137-B3.0
Page 72 of 81
paving material,
the relative
humidity shall be
at least 30%.
EFT
(GB/T 17626.4)
To power cable:
±2kV
To long I/O
cable: ±1kV
To power cable:
±2kV
To long I/O
cable: ±1kV
Power supply
grade shall be
minimally the
grade for typical
commercial or
medical
environment.
Surging
(GB/T 17626.5)
DM: ±1kV
CM: ±2kV
DM: ±1kV
CM: ±2kV
Power
frequency
magnetic field
(50/60Hz)
(GB/T 17626.8)
3A/m
3A/m
Power frequency
magnetic field
shall be of the
horizontal
characteristics as
in typical
commercial or
medical
environment.
Voltage sag,
short
interruption
and variation
(GB/T
17626.11)
5%UT (95%
fall, UT), 0.5
cycle;
5%UT (95%
fall, UT), 0.5 cycle;
Power supply
grade shall be
minimally the
grade for typical
commercial or
medical
environment. It is
recommended to
use UPS to
ensure
continuous
operation of this
product even in
case of AC power
outage.
40%UT(60% fall,
UT), 5 cycles;
40%UT(60% fall,
UT), 5 cycles;
70%U T (30%
fall, U T ), 25
cycles;
5% UT(95%
fall, UT), 5s;
70%U T (30% fall,
U T ), 25 cycles;
5% UT(95%
fall, UT), 5s;