9.ERROR CODE
98
Sequence Error (Error Assistance Code =3)
* However, SP1 Final SP1
9-2. Error Description
9-2-1. Calibration Error
Calibration Error 1
Zero calibration must be performed again. In standard calibration procedures,
Zero calibration is performed first, followed by Span calibration. However, if the
result of the Span calibration is significantly off the target, F159 displays
“Calibration Error 1”. If this happens you must perform Zero calibration.
Correctly performed Zero calibration will clear the error message.
Name Description Error Code
Sequence Error 1
Stop signal had been raised when the system tried to start
weighing sequence
1
Sequence Error 2
Stop signal raised while a weighing sequence is underway
2
Sequence Error 3
Auto Zero is activated due to AZ countdown. The system
enters into ZALM state
3
Sequence Error 4
Near zero signal was OFF when the system tried to start
weighing sequence (depends on system setting)
4
Sequence Error 5
When the system tried to start a weighing sequence, the
initial weight was found larger than SP1 (depends on system
setting)
5