6.WEIGHING MODE SETTING AND OPERATION
76
Example) Final 20.000
FF CPS. regulation value 0.100
Count Of FF CPS 4
FF CPS. Coefficient 2 / 4
* When set value for FF CPS. is changed, the count setting for Free fall
compensation is cleared and, and its counter value is also cleared.
Count Measured value Error FF CPS. counter FF CPS.
00 ← Power ON
1 20.050 +0.050 1 0.500
2 20.040 +0.040 2 0.500
3 20.070 +0.070 3 0.500
4 20.080 +0.080 4 → 00.500
+0.240/4 = 0.060
0.060 × 2/4 = 0.030 →
Corrected value
5 20.020 +0.020 1 0.530
6 20.000 0.000 2 0.530
7 20.010 +0.010 3 0.530
8 20.110 (+0.110)←×30.530
9 20.010 +0.010 4 → 00.530
+0.040/4 = 0.010
0.010 × 2/4 = 0.005 →
Corrected value
10 19.880 (-0.120)←×10.535
11 19.990 -0.010 1 0.535
12 20.010 +0.010 2 0.535
13 20.000 0.000 3 0.535
14 19.980 -0.020 4 → 00.535
-0.020/4 = -0.005
-0.005 × 2/4 = -0.003 →
Corrected value
0.532
FF CPS. Coefficient calculation
The value for the coefficient can be selected from 1/4, 2/4, 3/4
or 1.
In situation where you can expect fairly consistent measured
values, the value 1 is usually sufficient. When measured
values fluctuate, selection of 1/4 or 2/4 is recommended for
better accuracy.