4.CALIBRATION
48
Calibration Error 1
Zero calibration must be performed again. In standard calibration procedures,
Zero calibration is performed first, followed by Span calibration. However, if the
result of the Span calibration is significantly off the target, F159 displays
“Calibration Error 1”. If this happens you must perform Zero calibration.
Correctly performed Zero calibration will clear the error message.
Calibration Error 4
Value set for Balance weight and/or Span calibration is larger than Capacity
setting. Modify setting for Balance weight and/or Span calibration and retry
Span calibration.
Capacity and Balance weight value
For accurate Span calibration, Balance weight value should be selected between
50%-100% of Capacity.
Calibration Error 5
Balance weight value is preset to “00000”. Select a proper non-zero value.
Calibration Error 6
The load cell output falls short of the F159’s allowable span adjustment range.
Check if the load cell is loaded properly, and if its output specification meets
F159 requirements. Then, perform Span calibration again.
Load
Capacity
50
0
Margin
(9Minimum scale division)
Balance weight value