4.CALIBRATION
49
Calibration Error 7
Load cell output varies in negative range. Check if the cell is loaded in opposite
direction, or +SIG and –SIG of the load cell are connected in reverse order.
Then, perform Span calibration again.
Calibration Error 8
Load cell output is outside the F159’s allowable span adjustment range. Check if
the load cell is properly loaded, and the its rated output falls within the F159’s
span adjustable range. Perform Span calibration again. Then, perform Span
calibration again.