29
7 OPTION
29
OPTION
Chapter
7
Obtaining current output signal
Use the CUR. and GND terminals of the
F331 for connection with an external
device (500 or less load resistance).
Setting of D/A zero and gain
With the D/A converter of the F331, an analog
output is obtained by setting the indicated value
to output 4mA (D/A zero scale) and the
indicated value to output 20mA (D/A full
scale). Respective set values are inputted by the
D/A zero and full scale setting functions.
Example of setting
In the case where:
D/A output mode...................... 0 (linked with indicated value)
D/A zero scale.......................... 0500
D/A full scale ........................... 2500
CAUTION
- The D/A converter current output is an option.
- Do not apply external currents. Breakage will result.
Also, connecting a capacity load may cause oscillation.
External device
Load resistance;
500 or less
FG
Inside Outside
CUR.
GND
F331
Current (mA)
20
4
D/A zero scale
Indicated
D/A full scale
Resolution 1/10000
value
Current (mA)
20
4
D/A zero scale
Indicated value
D/A full scale
Resolution 1/10000
500 2500
Indicated value Current (mA)
480 3.84
500 4.00
1000 8.00
1500 12.00
2500 20.00
2520 20.16
Zero scale