Chapter 4 – Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 111/168
Learning cycles:
"Good part" learning cycle: this cycle is
compulsory before control tests.
This P1 and P2 learning cycle
automatically calculates and identifies the
value of the good part together with the
maximum and minimum limits (+/- 5 %
modifiable).
Fill Trans Time
P1
P2
Pressure
Max. limit (+5 %)
Min. limit (-5%)
"Bad part" learning cycle: this cycle is
not compulsory. It calculates the actual
minimum limit in relation to a bad part by
taking an average between the value of a
good part and that of a bad part
Fill Transfer Time
P1
Bad part
Pressure
Good part
Recalculated
limit
/2
The instrument allows for variations in input pressure. This is why the
maximum and minimum parameters vary for each cycle.
Learning cycles for good parts and parts with a large leak are accessed through the
main menu, special cycles. See paragraph 3.3.8. "Sealed component learning".
)Select the option and enter settings if necessary.
Test cycles can not be carried out unless learning cycles have previously been
carried out.