Chapter 6 – Error messages
UM-18400E-U ATEQ F580 User manual Page 156/168
2. ERROR FOR ONE HEAD
The faults are displayed for a head,
PROBLEM LIT INDICATORS MESSAGE DISPLAYED
Reference error. Leak in
excess of the full scale.
Action : check the
reference circuit.
T
R
CYCLE/P
r
001
PRESS=0.942 bar
LEAK=>>F.S REF.
READY (NO OK)
Test error. Leak in excess
of the full scale.
Action : check the test
circuit.
T
R
CYCLE/Pr001
PRESS=0.942 bar
LEAK=>>F.S TEST
READY (NO OK)
Pressure in excess of the
full scale.
Action : decrease the
pressure using the
mechanical regulator knob
or the target if an electronic
regulator is used.
T
R
CYCLE/P
r
001
PRESS=> FULL SCALE
READY (NO OK)
Differential transducer fault.
Action : contact ATEQ
after sales service for repair
(probably water or oil in the
instrument's test circuit).
T
R
CYCLE/P
r
001
PRESS=0.942 bar
LEAK= SENSOR DEF.
READY (NO OK)
Pressure in excess of the
max. threshold.
Action : check regulator
settings, pressure limits,
check whether the right
regulator has been selected
if there are two.
T
R
CYCLE/Pr001
PRESS=1.02 bar P>
READY (NO OK)
Pressure below the min.
threshold.
Action : check the network
pressure and regulator
settings, the pressure limits,
and whether the right
regulator has been selected
if there are two.
T
R
CYCLE/Pr001
PRESS=0.000 bar P<
READY (NO OK)