Chapter 6 – Error messages
UM-18400E-U ATEQ F580 User manual Page 157/168
PROBLEM LIT INDICATORS MESSAGE DISPLAYED
ATR error.
Action : run another ATR
learning cycle or check the
ATR parameters. ATR
fault.
T
PRESS=0.942 bar
LEAK=ATR DEFAULT
READY (NO OK)
Fault: no CAL learning or
CAL drift found following
request for CAL check.
Action : carry out CAL
learning or check the
programmed percentage of
CAL drift, the known
standard leak and the test
pressure.
T
PRESS=0.942 bar
LEAK=CAL DEFAULT
READY (NO OK)
CAL drift, following a CAL
checking request.
Action : check the rate
programmed for the CAL
drift, check the master leak,
and check the test
pressure.
T
PRESS=0.942 bar
LEAK=CAL DRIFT
READY (NO OK)
Commutation fault in the
equalising valve.
Action : check network
pressure, contact ATEQ
after sales service for
repair.
T
PRESS=0.942 bar
LEAK=VALVE DEF.
READY (NO OK)
1) The electronic regulator
has been unable to initialise
correctly.
2) The input pressure on
the regulator must be at
least 10 % of regulator full
scale + 100kPa (+ 1 bar).
Action : check supply
network pressure or
pressure at the regulator
input.
T
The indicator lights come on
briefly after the following
message is displayed :
REGULATOR
ERROR