Preamble
UM-18400E-U ATEQ F580 User manual Page 6/168
2. MEASUREMENT CHARACTERISTICS
2.1. PRESSURE DROP MEASUREMENT
RANGE
ACCURACY
Maximum
RESOLUTION
0 – 50 Pa +/- (2,5% of the pressure + 1 Pa 0,01 Pa
0 – 500 Pa +/- (2,5% of the pressure + 1 Pa 0,1 Pa
0 – 5000 Pa +/- (2,5% of the pressure + 10 Pa 1 Pa
2.2. TEST PRESSURE MEASUREMENT
RANGE
ACCURACY
Maximum
RESOLUTION
PE = 75 mbar* +/- (1,5% of the pressure+ 0,2 hPa) 0,1 % PE
PE < 0,3 bar +/- (1,5% of the pressure + 1 hPa) 0,1 % PE
0,3 PE 1 bar
+/- (1,5% of the pressure + 3 hPa) 0,1 % PE
1 < PE 5 bar
+/- (1,5% of the pressure + 7.5 hPa) 0,1 % PE
5 < PE 10 bar
+/- (1,5% of the pressure + 15 hPa) 0,1 % PE
10 < PE 20 bar
+/- (1,5% of the pressure + 30 hPa) 0,1 % PE
20 < PE 50 bar
+/- (1,5% of the pressure + 75 hPa) 0,1 % PE
PE = 100 bar
+/- (1,5% of the pressure + 150 hPa) 0,1 % PE
* Specific (relative).
2.3. P
RESSURE REGULATION
Mechanical regulator Electronic regulator
- 80 kPa to - 2 kPa
5 kPa to 50 kPa
20 kPa to 400 kPa
50 kPa to 800 kPa
50 kPa to 1600 kPa
- 80 kPa to - 2 kPa
1 kPa to 10 kPa
5 kPa to 50 kPa
20 kPa to200 kPa
50 kPa to 500 kPa
100 kPa to 1000 kPa
100 kPa to 1600 kPa
100 kPa to 2000 kPa
For other specific pressures please contact ATEQ.