Chapter 4 – Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 116/168
b) Setting the parameters
Validate the function in the
CONFIGURATION/EXTENDED
MENUS menu.
CONFIG
RECOVERY : NO
SEALED COMPONENTS : NO
REFERENCE VOLUME : NO
VOLUME CALCULATION : NO
STAMPING : NO
PEAK HOLD : NO
TEMP CORRECTION 1 : NO
INDIRECT : YES
SIGN : NO
11:26:18
01/07/2002
Validate by "YES" the function in the
PARAMETERS/HEAD XX/PRYY/
FUNCTION menu and enter the value
of the pressure supply monitoring limits
in the part (minimum and maximum fill).
The
AZ PIEZO 2 function enables the
setting up of and auto zero (reset) on
the second piezo sensor.
PARAMETERS
INDIRECT : YES
AZ PIEZO 2 : NO
Max FILL : 0.000 bar
Min FILL : 0.000 bar
11
:
2
9
:
1
3
0
1/07/2002
Note: for the Piezo 2 auto zero function, if the high pressure sensor is not fitted with an
Auto Zero valve, this parameter must be set to "NO".
The pressure monitoring limits for the
bell must be entered in the test
parameters menu.
PARAMETERS
AUTO PARAMETRING : NO
COUPLING A : 02.0 s
FILL : 01.0 s
STABILISATION : 00.5 s
TEST : 01.0 s
DUMP : 00.5 s
PRESSURE UNIT : bar
Max. FILL : 10.0 bar
Min. FILL : 0.000 bar
15:28:34 20/08/2002
Note: the pressurisation (low pressure) of the bell, enables the monitoring of the
integrity of the measurement circuit and bell.
During the fill and stabilisation phases,
the two pressures are displayed.
First line, pressure applied to the part,
second line, pressure in the bell (on the
above example, on the head 1).
ATEQ F580
---- H01 ----
RUN/Pr:001
PRESS = 1.50 bar
PRESS = 0.200 mbar
STABILISATION
---- H02 ----
RUN/Pr:001
PRESS = 2.00 bar
READY