EasyManua.ls Logo

ATEQ F580 - Page 132

ATEQ F580
180 pages
To Next Page IconTo Next Page
To Next Page IconTo Next Page
To Previous Page IconTo Previous Page
To Previous Page IconTo Previous Page
Loading...
Chapter 4 – Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 122/168
a) Parameters printout frame example
1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40
V e r s i o n 0 1 . 1 1 h
1 7 / 0 4 / 2 0 0 1
P r 0 1
T Y P E : L EA K
C O U P L I N G A : 0 0 . 1 s
F I L L . : 01 . 0 s
S T A B . : 01 . 0 s
T E S T : 0 2. 0 s
D U M P : 0 1. 0 s
M a x . F I L L : 2 . 4 0
M i n . F I L L : 1 . 6 0
T E S T F A I L : 0 1 0
R E F . F A I L : 000
Note: The "PROGRAM NAME" characters are printed when a program name has been
set in the parameters.
2.2.9. RS485
This function enables the configuration of the RS485 output to a C540 central or F580
when the instrument is installed in a network.
) Select the option and enter settings if necessary.
2.2.10. Modbus
This function enables the configuration of the Modbus link when the instrument is
installed in this type of network. The frame parameters, RS parameters (speed, serial
port) must be entered.
) Select the option and enter settings if necessary.
2.2.11. Security
This function deactivates the START and RESET keys on the instrument front panel.
Programs can only be started from the instrument inputs (J3 connector).
) Select the option and enter settings if necessary.

Table of Contents

Related product manuals