F9 Fetal & maternal Monitor Service Manual Functional Checks
- 13 -
If you touch the center of a key but the monitor does not execute the right operation, calibrate the
touch screen. Refer to section 5.5 Calibrating Touch Screen for details.
If any operation failure is detected, refer to section 8.3 Touch Screen Failures for details.
4.5 Printing Check
Press the PRINT key to start printing. Check if the recorder starts printing.
Enable the Print Self-Test function in the recorder’s configuration. Restart the monitor and verify
if the recorder prints a baseline.
Enter DEMO mode and print some traces and check if all the traces are clear on the paper.
If any failure is detected, refer to section 8.5 Recorder Failures for details.
NOTE:
Please make sure the paper loading and setting are correct before the printing starts.
4.6 Alarms Check
Stimulate a signal that is higher than the upper limit or lower than the lower limit to activate a
physical alarm. Disconnect one of the plugs to of the monitor to activate a technical alarm. Check
if the audible alarms and visible alarms are working properly.
If any failure is detected, refer to section 8.7 Alarm Failures for defective details.