Chapter 4 Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 135/168
3.3.6. 2) CAL check
This special cycle is used to check calibration in calibrated mode. See the explanation
in the previous section. The CAL check cycle checks whether calibration has drifted
beyond the limits set as a percentage. If the limits have been exceeded, an alarm is
triggered and a calibration cycle or check of the instrument will be necessary.
Amongst the special cycles available
select the "CAL check" cycle.
RUN
RUN PROGRAMME : Pr01 LEAK TEST
SPECIAL CYCLE : Inactif
15:14:31 06/11/2002
Regul. 1 adjustment
Regul. 2 adjustment
Infinite fill
Piezo auto zero
CAL learning
CAL check
Next, enter the instruction value, and
enter the reject unit.
Then start the special cycle by using
the START key.
RUN
RUN PROGRAMME : Pr:01 LEAK TEST
SPECIAL CYCLE : CAL check
1
5
:
2
5
:
2
5
06
/
11
/2002
At the end of the cycle, the display
gives the test result in the calibration
unit.
3.3.6. 3) Custom unit Check + Custom unit learn
This special cycle enables the checking of learning in the custom unit mode (Please
refer to the previous paragraphs for more information). The checking cycle measures
the drift in relation to the imposed percentage limits. If the limits are not exceeded, a
Custom Unit learn cycle will be carried out automatically to refresh the value learnt.
If the percentage limit is exceeded, the instrument will display a Custom Unit drift error.
The custom unit check + Custom unit learn cycle request can be carried out on input 7
of the J3 which has been programmed for this function or on a pre programmed key on
the optional RC5 consol if it is installed.