Chapter 4 Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 140/168
3.4.1. 2) Differential (Leak) sensor
This procedure concerns the calibration of the differential sensor. Before carrying out
this operation the pressure should be set to 0 on the mechanical regulator.
Select the SPE CYCLE menu. Then
select the LEAK SENSOR CHECK
cycle.
CYCLE
RUN PROGRAM : Pr01 LEAK TEST
SPECIAL CYCLE : Disable
09:58:31 03/10/2003
Disable
CAL. P1 REG1
CAL. P1 REG2
CAL. PRESSURE 2
LEAK SENSOR CHECK
VALVE AUTO TEST
When the cycle is selected, press on
the CYCLE START key to launch the
calibration cycle.
The instrument carries out a dump,
checks that the pressure is not zero
then switches to test and displays the
pressure measured. A sensor
calibration is then possible.
ATEQ F580
---- H01 ----
RUN/Pr:001
PRESS = 1.50 bar
TEST
---- H02 ----
RUN/Pr:001
PRESS = 2.00 bar
READY
Stop the special cycle by pressing on
the END OF CYCLE key.
ATEQ F580
---- H01 ----
RUN/Pr:001
PRESS = 1.50 bar
READY
---- H02 ----
RUN/Pr:001
PRESS = 2.00 bar
READY