Chapter 4 Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 141/168
3.4.2. Valve auto-test
Le cycle VALVE AUTO-TEST must only be carried out with plugs on the test and
reference outputs.
The test parameters are automatically attributed according to the active program (test
pressure) and the characteristics of the instrument.
To carry out an Auto-test cycle, select
the SPE CYCLE function in the main
menu, then the AUTO TEST function
and confirm.
CYCLE
RUN PROGRAM : Pr01 LEAK TEST
SPECIAL CYCLE : Disable
09:58:31 03/10/2003
Disable
CAL. P1 REG1
CAL. P1 REG2
CAL. PRESSURE 2
LEAK SENSOR CHECK
VALVE AUTO TEST
When the cycle is selected, press on
CYCLE START to launch the
calibration cycle.
During the Auto-test, the instrument
displays the pressure measured.
ATEQ F580
---- H01 ----
RUN/Pr:001
PRESS = 1.50 bar
AUTO-TEST
FILL
---- H02 ----
RUN/Pr:001
PRESS = 2.00 bar
READY
The cycle stops automatically if the
valve has no defect, the instrument
displays PART PASS.
ATEQ F580
---- H01 ----
RUN/Pr:001
PRESS = 1.50 bar
READY
---- H02 ----
RUN/Pr:001
PRESS = 2.00 bar
READY