Chapter 5 - Accessories
UM-18400E-U ATEQ F580 User manual Page 147/168
2.3.2. Electrical connections
2.3.2. 1) I/O inputs J1 connector (TOR)
1 2 3 4 5 6 7 8 9 10111213141516
VOLT FREE CONTACTS OR CNOMO INPUTS DEPENDING ON CONFIGURATION
Pin Extended mode Standard mode Compact mode
1 COMMON (outputs 2 to 8) COMMON (outputs 2 to 8) COMMON (outputs 2 to 8)
2 Pass Head 1 Pass Head 1 Pass
3 Test part fail Head 1 Fail Head 2 Pass
4 Reference part fail Head 1 alarm Head 3 Pass
5 Alarm Head 1 end of cycle
Bloc 1
Head 4 Pass
6
Pressure regulator fault
(pressure outside limits)
Head 2 Pass Head 5 Pass
7 End of cycle Head 2 Fail Head 6 Pass
8 Recoverable parts Head 2 alarm
Bloc 2
Head 7 Pass
9
COMMON (outputs 10 to
16)
COMMON (outputs 10 to
16)
COMMON (outputs 10 to
16)
10 Customer calibration fault Head 2 end of cycle Bloc 2 Head 8 Pass
11 Auto-diagnostic fault Head 3 Pass Head 9 Pass
12 ATR fault Head 3 Fail Reserved
13 Free Head 3 alarm Reserved
14 Free Head 3 end of cycle
Bloc 3
Reserved
15 Free Global end of cycle Global end of cycle
16 Free Special cycle fault Alarm
Refer to the diagram Chapter 1 paragraph 3.3.13. "J8 connector inputs planner" which
takes on the same conditions as for the connector J1 of the I/O module.
2.3.2. 2) J2 step by step push button
Button for numbering the head in the network, refer to
Chapter 3 paragraph 2 "start-up of the network".
2.3.2. 3) J3 input connector (RS485)
2 1
4
3
Reserved for ATEQ network.
Used for connection to other ATEQ instruments.
(Lumberg type male connector).
Pin 1 Network (D+) Pin 3 Network (D-)
Pin 2 + 24 V voltage Pin 4 0V earth
2.3.2. 4) J4 output connector (RS485)
2
1
4
3
Reserved for ATEQ network.
Used to connect to remote heads. 2 meter, 5 meter or 10 meter
cables are supplied depending on the configuration of the
installation.
(Lumberg type female connector).
Pin 1 Network (D+) Pin 3 Network (D-)
Pin 2 + 24 V voltage Pin 4 0V earth