Chapter 1 – Installation of the instrument
UM-18400E-U ATEQ F580 User manual Page 26/168
b) PLC in PNP mode connection
I
1
(Reset)
Customer
Inputs board
16/32 programs
I
2
(Start)
I
3
(Pr 1 + 1)
I
4
(Pr 2 + 1)
Good part
Common
Test fail part
Reference fail part
Alarm
End of cycle
24 V DC
24 V DC
ATEQ or customer
I
6
(Pr 8 + 1)
I
5
(Pr 4 + 1)
I
7
(Programmable input)
16
15
14
13
12
11
10
9
8
7
6
5
4
3
2
1
PNP MODE
Ground
+ 24 V DC
J3
INPUTS OUTPUTS
OUTPUTS
+
-
INPUTS
Customer
0 V
(0,3 A max)
Note: The 24V power supply
must be provided by the
internal power supply of the
ATEQ instrument (0,3A
maximum) OR through an
external power supply provided
by the customer.
In the case of customer
external supply, the ATEQ
instrument can be supply by
the 2 and 4 pins on the J3
connector too.
c) Lights connection
I
1
(Reset)
Customer
Inputs board
16/32 programs
I
2
(Start)
I
3
(Pr 1 + 1)
I
4
(Pr 2 + 1)
Good part
Common
Test fail part
Reference fail part
Alarm
End of cycle
24 V DC
I
6
(Pr 8 + 1)
I
5
(Pr 4 + 1)
I
7
(Programmable input)
16
15
14
13
12
11
10
9
8
7
6
5
4
3
2
1
Ground
+ 24 V DC
J3
INPUTS OUTPUTS
OUTPUTS
0,3 A Max
with ATEQ
24 V DC
!
0,2 A
Max
0,2 A
Max
(0,3 A max)
Note: The 24V power supply
must be provided by the
internal power supply of the
ATEQ instrument (0,3A
maximum) OR
through an
external power supply
provided by the customer.
In the case of customer
external supply, the ATEQ
instrument can be supply by
the 2 and 4 pins on the J3
connector too.