Chapter 4 Functions of the instrument
UM-18400E-U ATEQ F580 User manual Page 131/168
3.3.3. Piezo reset
Used for compulsory reset of the piezo transducer.
Select amongst the special cycles
available the "Piezo auto zero" cycle.
Then, start the special cycle by using
the START key.
CYCLE
RUN PROGRAMME : Pr01 LEAK TEST
SPECIAL CYCLE : Disabled
17:45:18 06/11/2002
Disabled
Regul. 1 adjustment
Regul. 2 adjustment
Infinite fill
Piezo auto zero
At the end of the special cycle. The
instrument stops its cycle and
configures itself for a new measurement
test.
ATEQ F580
---- H01 ----
RUN/Pr:001
PRESS = 1.50 bar
READY
---- H02 ----
RUN/Pr:001
PRESS = 2.00 bar
READY
Note: the auto zero cycle is an automatic cycle which is carried out approximately every
5 minutes. It is used to initialise the pressure transducers in relation to atmospheric
pressure.
For automatic devices it may be necessary to inhibit the auto zero cycle.
To disable the automatic auto zero
cycle, enter the "CONFIGURATION"
menu, select the "PIEZO AUTO ZERO"
function and confirm with "NO"
CONFIG
LANGUAGE : English
AUTO SAVE : YES
PIEZO AUTO ZERO : YES
PRINTER : NO
SECURITY : NO
I/O GROUP : 1
EXTERNAL DUMP : NO
LIGHT
Attention: If an auto zero cycle is not carried out regularly, measurement errors may
occur and result in false readings for the air tightness of parts.