EasyManua.ls Logo

Unipulse f395 - 23 .Specifications; Analog Input Section

Unipulse f395
180 pages
To Next Page IconTo Next Page
To Next Page IconTo Next Page
To Previous Page IconTo Previous Page
To Previous Page IconTo Previous Page
Loading...
23.Specifications
155
23. Specifications
23-1. Analog Input Section
Two-system inputs
(Standard: CH1=strain gauge input, CH2=voltage input)
(Optionally, changeable as follows:)
Strain gauge input
Sensor excitation voltage: Switching between 10V and 2.5V DC (by setting)
Output current: 120mA or less (total of two inputs)
Four-wire type (When only one input is used, up to
four 350Ω strain gauge type sensors can be connected
to the F395 in parallel.)
Signal input range: -3.0mV/V ~ +3.0mV/V
Zero/gain adjustment: By digital operation.
Equivalent input calibration range:
-3.0mV/V ~ -0.5mV/V
+0.5mV/V ~ +3.0mV/V
Equivalent input calibration range error:
Within 0.2% F.S.
Accuracy: Non-linearity: Within 0.02% F.S. ± 1 digit
(at 3.0mV/V input).
Zero drift: Within 0.5μV/ RTI
(input-converted value).
Gain drift: Within 0.01%/ .
Analog filter: Selectable from 10, 50, 200, and 600Hz
A/D converter: Speed: 4000 times/sec. (max.) when one input is used.
2000 times/sec. (max., X-axis measurement)
when two inputs are used.
Changeable to
100, 200, 500, 1000, 2000 times/sec.
according to the input waveform.
Resolution:
Approx. 1/30000 relative to 3.0mV/V
Model CH1 CH2
VI1 Voltage Voltage
CI1 Current Voltage
LI2 Strain gauge
(Load cell)
Strain gauge
(Load cell)
CI2 Strain gauge
(Load cell)
Current

Table of Contents