4.Methods of Connection
22
4-1-1-2. Strain Gauge Sensor Connection
Before connecting a sensor, set the excitation voltage, turn off the power, and perform
the following connection. (See "5-6. Excitation Voltage" P34.)
4-wire sensor
6-wire sensor
For connecting a 6-wire strain gauge type sensor, short-circuit +EXC and +S, and -EXC
and -S.
For connecting a sensor, be sure to use shielded lines, and separate
wiring from lines with heavy noise (such as wiring of power equipment
and wiring of digital equipment) and AC lines.
The analog GND in the F395 is grounded to F.G. Also, when voltage/
current input is specified, the -SIG terminal is at about the same level
as the analog GND. (See "20. Block Diagram" P152.) Carry out wiring
so that unnecessary current will not flow.
Request
EXC
EXC
SIG
SIG
F.G
IN
OUT
IN
OUT
F395
EXC
EXC
SIG
SIG
F.G
IN
OUT
IN
OUT
S
S
F395