5.Methods of Calibration
36
5-11. Equivalent Input Calibration
Set the rated output value of the sensor.
1. Select the "Equivalent Input Calibration" item.
2. Input set values as follows:
 ("i.- iii." depend on the analog input specifications (factory-shipped settings).)
i. Strain gauge type sensor
Sensor Output mV/V? .. Setting range: -9.999 ~ -0.100, 0.100 ~9.999
Load? ............................ Setting range: 100 ~ 9999 (Not including the decimal point.)
ii. Voltage output sensor
Sensor Output V?.......... Setting range: -9.999 ~ -0.100, 0.100 ~ 9.999
Load? ............................ Setting range: 100 ~ 9999 (Not including the decimal point.)
iii. Current output sensor
Sensor Output mA? ...... Setting range: -99.99 ~ -1.00, 1.00 ~ 99.99
Load? ............................ Setting range:100 ~ 9999 (Not including the decimal point.)
3. For error display, see "19. Error Messages" P150.
MODE CAL. Equiv. CAL.