5.Methods of Calibration
32
5-3. Equivalent Input Calibration Procedures
Perform equivalent input calibration of the sensor as follows:
Turn off the LOCK switch on the rear panel to unprotect
calibration.
Set the "Set Value Protection" setting to unprotect
calibration.
Select the "Excitation Voltage" from 2.5V/10V.
(Set 2.5V when the input specification is voltage / current.)
Turn off the power, and connect the sensor.
Set the "Unit" to be used.
Set the "Decimal Place".
Set the minimum value of the digital changes.
Set the initial zero point.
Set the rated output value of the sensor.
Protect calibration to prevent misoperation.
Turn on the LOCK switch to protect calibration to prevent
misoperation.