5.Methods of Calibration
34
5-5. LOCK Release
Calibrated values and set values can be locked to prevent a change due to misoperation.
There are two types of locks: a software lock to be operated from the screen, and a
hardware lock by the switch on the rear panel. For performing calibration, release both
locks.
1. Turn off the LOCK switch on the rear panel. (Hardware lock)
2. Set the "Set Value Protection" setting to "All Unprotect". (Software lock)
Now, the locks are released. Upon completion of calibration, hardware lock and software
lock again to prevent misoperation.
5-6. Excitation Voltage
Set the voltage value to be applied to the bridge of the strain gauge sensor. Select as large
a value as possible within the range not exceeding the maximum excitation voltage of the
sensor.
Setting item: 2.5V / 10V
MODE System Protect All Unprotect
MODE CAL. EXC. Volt
CH1
CH2