EasyManuals Logo

Unipulse F325 User Manual

Unipulse F325
160 pages
To Next Page IconTo Next Page
To Next Page IconTo Next Page
To Previous Page IconTo Previous Page
To Previous Page IconTo Previous Page
Page #88 background imageLoading...
Page #88 background image
77
12 Interface
77
Interface
Chapter
12
Connection of RS-485
Two-wire (point to point)
- Use twisted pair wires for connection cables. (Noise margin increases.) However, two-core
parallel cables are sufficient for short-distance connection.
- Install the included terminator to the F325 if the F325 is a dead end.
- SG terminal (B10 terminal) is a ground terminal (which protects circuits) used on the circuit.
SG terminal is not normally required to use if the main unit of the F325 and connection
counterpart device are class D grounded.
However, in case connection is necessary based on the on-site conditions, check the
specifications of the counterpart device and connect it.
Two-wire (multi point)
A-
B+
A
B
A
B
A-
B+
F325
PLC
programmable display unit and so on
Terminator
(Master) (Slave)
* Logic
- Mark state (OFF)
V
A
V
B
<- 0.2V
- Space state (ON)
V
A
V
B
0.2V
V
A
 A- terminal voltage
V
B
 B+ terminal voltage
Terminator
(accessory)
Depending on master devices, A and B may be reversely indicated.
In case that communication is not possible, switch A and B.
Attention
A
B
A- B+ A- B+ A- B+ A- B+
Twisted pair wire
F325 F325 F325 F325
Terminator
Terminator

Table of Contents

Questions and Answers:

Question and Answer IconNeed help?

Do you have a question about the Unipulse F325 and is the answer not in the manual?

Unipulse F325 Specifications

General IconGeneral
BrandUnipulse
ModelF325
CategoryMeasuring Instruments
LanguageEnglish

Related product manuals