4.CALIBRATION
16
4. CALIBRATION
Calibration is performed for matching the F371 to a strain-gage sensor. The following
two types of calibration are available for the F371.
Equivalent Input Calibration
Calibration is performed without an actual load by entering the rated output value (mV/
V) and the capacity (to be indicated) of the strain-gage sensor by the keys. Calibration is
easily performed when no actual load is available.
For example, the gain is automatically determined by entering:
2.001mV/V (rated output) - 100.0kg (capacity)
as indicated for a load.
F371
Strain gauge sensor
Rated output value (mV/V)
Indicated value
A data sheet is attached to a strain-gage sensor at the time of purchase.
The data sheet provides data including:
capacity load (in kg, t, etc.)
rated output voltage (in mV/V)
non-linearity, hysteresis, input resistance, output resistance
and zero balance.
Enter the capacity and the rated output value required for equivalent
input calibration into the F371.