Tip Calibration Manual Tip Calibration
Page 139
© Copyright 2020 Stratasys. All rights reserved.
c. Take the average value of the four measurements. This is the number you will enter for
the Z offset adjustment.
Figure 16: Remove and Measure Support Layer
d. If the value measured in s
tep c above is within ±0.0005 in. (0.01 mm) of the model tip’s
slice height of 0.010 in. (0.254 mm), the printer is calibrated for the Z axis and an
ad
justment is not required. Proceed to step 4.
e. If the value measured in s
tep c above is not within ±0.0005 in. (0.01 mm) of the model
tip’s slice height you will need to enter a Z offset adjustment using the Up
and Down
buttons within the Manual Tip Calibration page. Each button press equates to one ten
thousandths of an inch.
Peel support
layer here
Measure support layer
here on all 4 sides